Leon,
Assuming you are running HP Pascal 3.22 in a BD Consort 32 config-
uration, the issue is in the file Owner/Group specification after ftp'ing
back to the HP from another system. For example, take the STARCOMP.CODE
in its installation directory:
-type- --code-- ---Last Change--- -----Status------ --size-- ------name------
Code -5582 10-09-91 16:24:44 666m 17u 9g 156160 STARCOMP.CODE
Data -5622 06-07-94 17:54:26 666m 17u 9g 42 ACF_IDS
Data -5622 10-09-91 16:24:50 666m 17u 9g 212 ACF_CV
Data -5622 10-09-91 16:24:53 666m 17u 9g 160 CALIBC_F1
Data -5622 10-29-93 16:41:18 666m 17u 9g 160 CALIBS_F1
Data -5622 10-09-91 16:24:59 666m 17u 9g 3072 ADDAPLCNFG
Note the Owner/Group specification (an integer 0 - 65535) is -5582 for the
CODE file, and -5622 for the others. After emulating a "restore" over
ethernet via ftp (binary - FACSNet transfer):
-type- --code-- ---Last Change--- -----Status------ --size-- ------name------
Hp-ux -5813 11-08-94 16:09:02 666m 17u 9g 3072 ADDAPLCNFG
Hp-ux -5813 11-08-94 16:09:20 666m 17u 9g 42 ACF_IDS
Hp-ux -5813 11-08-94 16:09:36 666m 17u 9g 212 ACF_CV
Hp-ux -5813 11-08-94 16:10:08 666m 17u 9g 156160 STARCOMP.CODE
Note the Owner/Group specification gets laid down as -5813 by the Fusion
product (FACSNet). Now, you'd think that you could easily use the (horrid)
little Filer utility and the Hfs command to hack this number back to -5582
somehow (Consort System 32 User's Guide, 5-41 -> 5-43), but I can't see how
this number can be modified with Hfs. What we need to know is: Under Pascal,
how do we do the Unix equivalent of 'chmod a+x <filename>' to restore
execute permissions on a file. Maybe one of BD's Pascal guys could help?
Joe
![]() |
![]() |
![]() |
![]() |
![]() |
CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the
Purdue University Cytometry Laboratories and distributed free of charge
as an educational service to the cytometry community.
If you have any comments please direct them to
Dr. J. Paul Robinson, Professor & Director,
PUCL, Purdue University, West Lafayette, IN 47907.
Phone: (765)-494-0757;
FAX(765) 494-0517;
Web