>The formulae I use on our FACStarPLUS to calculate sort rate & "expected"
recovery (in a spread sheet) are:
R = (ma)*exp(-mnT)
where R = sort rate
m = flow rate
a = percent positive
n = number of drops deflected
T = 1/Period (@40 microseconds with a 70 micron nozzle)
exp = exponent ( 10^(-mnT) will work too)
"Expected" recovery is
P = exp(-(-1-a)uTn)
where u = frequency
At 1,200 chromosomes/s, 1% pure, 2 drops, frequency of 25,000; sort rate will be
@ 10/s with 14% recovery.
You'd need to start with over 4 x 10^8 cells!!!
I should also mention that I've only sorted whole cells!
Dennis
*****************************************************************************
* *
* Dennis J. Young Voice : (619) 543-3928 *
* Flow Cytometry Core Facility FAX : (619) 543-7487 *
* University of California, San Diego e-mail: djyoung@ucsd.edu *
*****************************************************************************
>-- Saved internet headers (useful for debugging)
>Received: from flowcyt.cyto.purdue.edu by ucsd.edu; id PAA28530 sendmail 8.6.9/
>Received: by flowcyt.cyto.purdue.edu (931110.SGI/930416.SGI.AUTO) id AA17059;
>Received: from uv1.im.med.umich.edu by flowcyt.cyto.purdue.edu via SMTP (931110
>Message-Id: <9411071809.AA12444@flowcyt.cyto.purdue.edu>
>Received: from IMVS4 by UV1.IM.MED.UMICH.EDU via MAIL-11 with DECnet; Mon, 7
>Date: Mon, 7 Nov 94 13:08:50 EST
>From: kweber@UV1.IM.MED.UMICH.EDU
>Subject: chromosome yield
>To: cytometry@flowcyt.cyto.purdue.edu
>X-CCaddr: Bcc: yashpal agrawal AT cis
>X-CCaddr: Bcc: dennis young AT cis
>X-CCaddr: *To: cytometry@flowcy
![]() |
![]() |
![]() |
![]() |
![]() |
CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the
Purdue University Cytometry Laboratories and distributed free of charge
as an educational service to the cytometry community.
If you have any comments please direct them to
Dr. J. Paul Robinson, Professor & Director,
PUCL, Purdue University, West Lafayette, IN 47907.
Phone: (765)-494-0757;
FAX(765) 494-0517;
Web