> Dear Paul/Steve;
>
> I am looking for references detailing a procedure for the titration of
> monoclonal antibodies used for two and/or three color staining of
> lymphocyte subsets in whole blood.
We do a 2- or 3-fold dilution series, bind these dilutions to cells for 15
min at room temp, rinse, (then FITC-anti-IgG 15 min on ice if the antibody
is not directly conjugated, rinse), optionally fix, then analyze.
Generally for monoclonals the results nicely fit the Michaelis-Menten
1-site binding equation. Using a program which fits the data to the
equation (we use Fig P from Biosoft, 314-524-8029), you get the
concentration which half-saturates (Ch) and the maximum intensity at
theoretical saturation (Imax) in an objective way which uses all data
points. The equation then lets you predict that e.g. at a concentration of
4Ch, you'll get 80% of saturation. The equation is simply I = Imax (C / (C
+ Ch)), where I is the intensity observed at concentration C. Note that
you must use the antibody CONCENTRATION not dilution, e.g. convert 1/1,000
to 0.1%. (I don't have a reference for this approach, which we arrived at
ourselves, but I'd like to know a reference if someone else has seen it in
print.)
/*- - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - -
Eric Martz, Professor of Immunology emartz@microbio.umass.edu
Dept Microbiology Voice: 413-545-2325 FAX: 413-545-1578
Morrill IVN 203, Box 35720, Univ Massachusetts, Amherst MA 01003-5720
- - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - -*/
![]() |
![]() |
![]() |
![]() |
![]() |
CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the
Purdue University Cytometry Laboratories and distributed free of charge
as an educational service to the cytometry community.
If you have any comments please direct them to
Dr. J. Paul Robinson, Professor & Director,
PUCL, Purdue University, West Lafayette, IN 47907.
Phone: (765)-494-0757;
FAX(765) 494-0517;
Web