Re: TIME on BD HP 310?

Mike_Lance@bdis.com
Wed, 31 Aug 94 10:31:16

Eric,

FACScan Reseach Software and FACStar Research Software do not record TIME as a
list mode parameter; LYSYS II is the only application from BD on the HP that
does generate time. LYSYS II can be run on a CONSORT 30 Color system.

For files generated by software that does not generate time separately, most
applications extrapolate time by using the $BTIM and $ETIM keyword values to
get the total time, then divide by the number of cells indicated in the $TOT
keyword value.

Best Regards,
Mike Lance

_______________________________________________________________________________
Subject: TIME on BD HP 310?
Author: emartz@microbio.umass.edu

Message-Id: <58382.emartz@microbio.umass.edu>
Date: 8/30/94 9:04 PM

Is there any way to record event acquisition TIME as a parameter using
FACScan or FACStar Research Software vintage 1988 on either a FACScan or
FACStar Plus? This software uses the old HP 310 computers, not the newer
HP 340's.

/*- - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - -
Eric Martz, Professor of Immunology emartz@microbio.umass.edu
Dept Microbiology Voice: 413-545-2325 FAX: 413-545-1578
Morrill IVN 203, Box 35720, Univ Massachusetts, Amherst MA 01003-5720
- - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - -*/


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu