> Does anyone happen to know the formula employed by Coulter for the
> calculation of "half-CV's" from flow histograms.
>
> Thanks in anticipation!
>
>
> David Galbraith
>
>
>
Hello David
To quote from the Coulter EPICS manual:
Half Cv= Nu-Ni
PC
i.e. (Nu-Ni)/PC
where Nu upper channnel having count equal to 0.5x the peak channel
count
Ni = lower channel having count equal to 0.5 x the peak channel count
PC = peak channel number
Nu Ni are determined using linear interpolation, but are always
within one channel of the channel nearest the peak on either side
having less than 1/2 the peak channel count. In order to use this
estimate of CV, the population must have a single peak and be smooth
enough so that the peak height and width at half width can be
accurately determined.
Best wishes,
Nigel Blackhall, Plant Genetic Manipulation Group, Dept Life
Science, University of Nottingham, Nottingham, UK
![]() |
![]() |
![]() |
![]() |
![]() |