> Does anyone out there have a protocol for using sodium borohydride? I
> want to try to use it to reduce autofluorescence (Cytometry
> 8:235-239(1987)).
I'll be very interested in your results. I tried using thiosulfate and
sulfite for chemical reduction following the paper you cite (and some
others I think) but got less than 2-fold drop in autofluorescence of
EBV LCL (human B cells, JY line) even with concentrations which clearly
did quite a bit of damage to the cell structure. An idea given to me
by Janos Szollosi is to prebleach the cells with an intense 488 nm
laser beam (spread enough with a lens to cover a test tube), but I
haven't had the guts (retinal fortitude?) to try that. We tried 15
min or so in direct sunlight but that also did almost nothing. Probably
the best idea is to use a spare photomultiplier (if you have one) to
measure autofluorescence in a channel not being used for a fluorochrome.
This can be used to subtract autofluorescence from other channels quite
nicely as mentioned by someone to this list recently. There are a number
of references available on variants of this technique. Searching MEDLINE
for autofluroescence is a good way to get them (back 10 years or so).
/*- - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - -
Eric Martz, Professor of Immunology emartz@microbio.umass.edu
Dept Microbiology Voice: 413-545-2325 FAX: 413-545-1578
Morrill IVN 203, Box 35720, Univ Massachusetts, Amherst MA 01003-5720
- - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - -*/
![]() |
![]() |
![]() |
![]() |
![]() |
CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the
Purdue University Cytometry Laboratories and distributed free of charge
as an educational service to the cytometry community.
If you have any comments please direct them to
Dr. J. Paul Robinson, Professor & Director,
PUCL, Purdue University, West Lafayette, IN 47907.
Phone: (765)-494-0757;
FAX(765) 494-0517;
Web