> I have been carrying out culture of CD34+ cells in the presence of
> IL-3, GM- & G-CSF and following myeloid differentiation using cell
> surface markers.
>
> I am finding that staining with CD14 is grossly under-estimating the
> monocytic content of these cultures as judged by morphology.
>
> One of my colleagues has made similar observations. Additionally he
> found that CD14+ AML cells lost expression of this marker when in
> culture.
>
> Has anyone had similar experiences or any good suggestions for
> alternative markers which more reliably indicate monocytic
> differentiation under these conditions?
>
>
CD14 will stain a subset only even of peripheral blood monocytes.
These cells are extremely heterogeneous and multi-color analysis
will reveal at least five subpopulations (Arterioscler Thromb Vasc
Cell Biol 16: 1437-47, 1996). To our knowledge there is no surface
marker available which will stain all monocytes. CD64 is similar
in expression to CD14 but is also expressed on activated
neutrophils. CD33 will detect more monocytes but all neutrophils
will also stain dimly. The ideal marker would be CD68 but I am not
aware of a good commercial clone which will react with the
extracellular portion of this antigen.
Gregor Rothe
*************************************************************************
Institute for Clinical Chemistry and Laboratory Medicine
University of Regensburg
D-93042 Regensburg, Germany
Tel. +49 (941) 944-6204
Fax +49 (941) 944-6202
Internet: Gregor.Rothe@klinik.uni-regensburg.de
*************************************************************************
![]() |
![]() |
![]() |
![]() |
![]() |
CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the
Purdue University Cytometry Laboratories and distributed free of charge
as an educational service to the cytometry community.
If you have any comments please direct them to
Dr. J. Paul Robinson, Professor & Director,
PUCL, Purdue University, West Lafayette, IN 47907.
Phone: (765)-494-0757;
FAX(765) 494-0517;
Web