Re: Wanted: extra BD sheath tank

Joseph Webster (J.Webster@centenary.usyd.edu.AU)
Mon, 17 Feb 1997 10:24:14 +1100

At 04:06 PM 14/02/97 -0400, Barry Grimes wrote:
>Hello all,
>I'm planning on doing sterile sorting and will be using ETOH to purge my
>FACSVantage. To avoid repeated filling and rinsing steps using only one
>tank, I would like to use an extra sheath tank for the alcohol rinse - but
>would rather not spend the $1200 for a spare 10 L tank. Is there anyone out
>there that might have a spare FACStar / Plus 2 L sheath tank they'd be
>willing to sell? Does anyone know if this will in fact work?

Yes it will certainly work, I've been doing it for years.
I use the waste tank for ETOH, and still use a transparent container
for waste.
(easy to see if it is about to overflow into the house vacuum line...)

Both tanks are online and pressurised, and I use a 3-way lauer-fit
valve to select 70% ETOH or saline.

NOTE:
Be very careful not to run the ETOH into the sheath tank, as the saline
will precipitate!

I also add an extra 1" 0.22u syringe filter into the line immediately
after running ETOH, which has several effects:
- Extra filtering of the saline, just in case.
- The small filter-full of air stops the saline & ETOH from mixing
in the line, so salt precipitation is not a problem.
- One minus, it changes the pressure balance so that the sample
differential must be reduced by about half with extra filter in.

I got this idea from Frank Battye of WEHI, with much appreciation!
Frank - you may have some comments; are you still using clear filter
holders as fluid reservoirs?

Cheers, Joseph.

--
Joseph Webster
Flow Cytometry Facility
Centenary Institute


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu