Re: FACSort/PowerMac

Matthias Haury (mhaury@pasteur.fr)
Fri, 14 Feb 1997 10:26:17 +0100

Concerning the FACS-Mac connection problems:

Which version of Cellquest you are using (3.01) ? There are known problems
with the statusboard in the FACScan (don't know in the FACSsort ?), BDIS is
working on this problem, they might come out with a new EPROM chip that
should solve some of those problems.

However in our case restarting the FACScan as well as the computer solves
the problem in 99% of the cases immediately.

There are other possible reasons for the problem:
1. Make sure you use System 7.5.3 Rev 2.0 US, that the HD has no errors
using Disk First Aid or Norton Utilities,
2. Have no other extensions/inits installed than the system ones and the
BDIS inits... (if it's not the statusboard or the acquisition card I
suspect it's a extension conflict).
3. Check the cable as well as the connection of the card... use some
special products to clean the oxidation on the contacts of the cable.
4. Take out the statusboard on the FACSsort and make sure that all the
chips and the eprom are well inserted.
5. Does the problem persist if you desactivate the network ?
6. Make sure that you try all this without SCSI devices (optical disks)
attached

Well, if any of those doesn't solve the problem, you should ask BDIS to
change the Statusboard and the Acquisition card, as either one might have a
problem.

Also, I think Cellquest 3.1 should come out fairly soon, and that will
solve some of those problems...

Hope that helps...

Matthias

_____________________________________________________________________________
Matthias Haury Flowcytometry Dept Immunology Institut Pasteur
mhaury@pasteur.fr Tel: 33 (01) 40 61 31 29 Fax: 33 (01) 45 68 86 39
_____________________________________________________________________________


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu