Re: negative control for cytoplasmic C-mu or CD3

Jeffrey A. Clapper (clap@iastate.edu)
Fri, 31 Jan 1997 09:08:45 -0600

t 07:28 PM 12/16/94 -0500, you wrote:
>
>Dear Flow-ers,
>
>My colleague and I would like to set up a test for the expression of
>cytoplasmic IgM for differential diagnosis of B-ALL and would
>appreciate any comments /advices /references on the problems that we
>encountered, mainly how would one tell the difference between a surface IgM
>versus a cytoplasmic IgM? Can I somehow block the staining of surface IgM
>or can I quench the surface staining as is being done with phagocytosis
>tests? And is there any good controls for the tests?
>
>We are also evaluating the possibility of performing a cytoplamic CD3
>staining for diagnosis of T-ALL and, quiet possibly use it for monitoring
>of minimal residual disease. We have the same concerns for this
>test as we mentioned for cytoplasmic c-mu and would appreciate any comments
>/advices /references.
>
>
>Warm regards,
>Austin Lin
>Core Facility Laboratory-Flow Cytometer,
>Cheng Gung Medical College
>Tainan, Taiwan
>
Austin:

We have used a two-color method to stain both surface and cytoplasmic IgM in
mouse splenocytes.

1. Incubate with biotinylated-anti-IgM in PBS-0.5% BSA-0.01% Azide (PBA)

2. Wash

3. Fix in PBS with 1% ultra-pure formaldehyde (Polysciences).

4. Wash

5. Fix in 70% ethanol (OR perform all of the following in PBA with 0.5%
saponin)

6. Incubate with anti-IgM FITC and strepavidin-PE in PBA

7. Wash

8. Fix again in PBS with 1% formaldehyde

9. Analyze

The cells with surface IgM (and some cytoplasmic IgM) will be double-positive.

The cells with stictly cytoplasmic IgM will be only FITC-positive.

Jeff Clapper
Cell and Hybridoma Facility
Iowa State Univeristy
515-294-8504


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu