1. INTRODUCTION
Usually, the first step to perform cytometric evaluation of apoptosis is the separation of peripheral blood mononuclear cells (PBMC) through a density gradient. This procedure allows the removal of granulocytes and red blood cells, but can generate several drawbacks, among which we can remember: i) consistent waste of time, ii) necessity of manipulating dangerous biological samples, iii) selective loss of cell subpopulations, and iv) cell activation by mechanical stimulation.
With the aim to overcome these problems, we set up a technique able to evaluate PBMC spontaneous susceptibility to PCD in short term whole blood cultures. The proponed method consists of a modification of Nicoletti's technique, and is based on multiparametric FACS analysis of physical parameters, DNA content and CD15 expression of peripheral blood leucocytes cultured for 24 or 48 h in RPMI 1640 medium supplemented with fetal calf serum (FCS).
A threshold set on forward scatter together with an electronic window selective for events with CD15 negativity allows to focus the analysis on a cellular subset mainly made up of PBMC (Figs. 1-3).
After properly setting the analysis gate, it is possible to carry out DNA analysis in this selected cell population, defining as apoptotic the cells with a hypodiploid DNA content.
Multiparametric analysis shows that the forward scatter of cells with hypodiploid DNA content is clearly different from cells with diploid content (fig. 2). This behavior is well-known, and is related to the cell shrinkage induced by the apoptotic process
Finally, the result of the analyses can be expressed as percentage of apoptotic cells, or as apoptotic index, i. e. the ratio between vital and apoptotic cells.
2.1 Materials
2.2 Methodology
2) At 24 and 48 h respectively, 1 ml of cell suspension was drawn from the culture, washed with cold PBS, resuspended in 100 µl of PBS containing 20 µl of anti-CD15 FITC-conjugated MoAb, and incubated for 20 min at 4°C.
3) Two ml of hypotonic NH4Cl solution were then added, and cells incubated for 10 min at room temperature, washed in cold PBS and resuspended in 2 ml of cold 70% ethanol for at least 1 h.
4) Finally, cells were washed in cold PBS, and resuspended in 1 ml isotonic fluorochrome solution (PI 10mg/ml in PBS). Cells were analyzed after a minimum of 30 min of incubation in this solution. Flow cytometry was performed using a FACScan flow cytometer.
Briefly, this method consists of a first level, low cost and easy to perform technique able to explore the susceptibility to PCD of whole blood PBMC.
Instead of removing granulocytes and red blood cells by centrifugation through density gradient, this technique focuses the DNA analysis on a multiparametric defined cell population. Its rationale is based on a widely accepted point, namely that the extrusion of apoptotic bodies gives the apoptotic cell a hypodiploid content of DNA. Its main clinical application could be the exploration of susceptibiliy to apoptosis in PBMC from patients with autoimmune or viral diseases; nonetheless, an application in the evaluation of peripheral blast response to anticancer drugs is possible.
This technique can be easily improved in a dual laser system using a red-excitable dye (e.g. ToPro3) for DNA analysis, and introducing more blu-excitable phenotypic or metabolic probes for further characterization of subsets interested by the apoptotic process. Such modifications are currently in development in our Laboratory.
3.2 Critical parameters
No critical parameters were found with the described procedure.
3.3 Troubleshooting.
Lower FCS concentration can result in an increase of debris, making difficult the isolation of lymphocyte cluster on physical parameters.
3.4 Anticipated results
In a group of 40 healthy persons (blood donors), the mean apoptotic index was 0.04 (range 0.01-0.09) at 24 h, and 0,06 (range 0.03-0.15) at 48 h. A significative linear correlation (r=0.62) was found between values measured at 24 and at 48 h (data not shown).
3.5 Time considerations
This method is very simple and fast. The time needed for the setting up of cultures is negligible, and the time requested by sample staining and analysis is similar to an ordinary immunophenotypic FACS analysis. Moreover, when resuspended in ethanol, the sample can be stored for several days at +4°C without perceptible loss of fluorescence.
3.6 Key references.
2) Cossarizza, A., C. Mussini, N. Mongiardo, V. Borghi, G. Kalachnikova, B. De Rienzo, and C. Franceschi. 1997. Mitochondria alterations and dramatic tendency to undergo apoptosis in peripheral blood lymphocytes during acute HIV syndrome. AIDS. 11:19-26.
3) Emlen, W., J. A. Niebur, and R. Kadera. 1994 Accelerated in vitro apoptosis of lymphocytes from patients with systemic lupus erythematosus. J. Immunol. 152:3685:3692.
4) Endresen, P. C., P. S. Prytz, and J. Aarbakke. 1995. A new flow cytometric method for discrimination of apoptotic cells and detection of their cell cycle specificity through staining of F-actin and DNA. Cytometry. 20:162-171.
5) Gougeon, M.-L., H. Lecoeur, A. Dulioust, M.-G. Enouf, M. Crouvoisier, C. Goujard, T. Debord, and L. Montagnier. 1996. Programmed cell death in peripheral lymphocytes from HIV-infected persons. Increased susceptibility to apoptosis of CD4 and CD8 T cells correlates with lymphocyte activation and with disease progression. J. Immunol. 156:3509-3520.
6) Jaleco, A. C., M. J. Covas, and R. M. M. Victorino. 1994. Analysis of lymphocyte cell death and apoptosis in HIV-2-infected patients. Clin. Exp. Immunol. 98:185-189.
7) Lorenz, H.-M., M. Grünke, T. Hieronymus, M. Herrman, A. Kühnel, B. Manger, and J. R. Kalden. 1997. In vitro apoptosis and expression of apoptosis-related molecules in lymphocytes from patients with systemic lupus erythematosus and other autoimmune diseases. Arthritis Rheum. 40:306-317.
8) McCloskey, T. W., N. Oyaizu, M. Coronesi, and S. Pahwa. 1994. Use of a flow cytometric assay to quantitate apoptosis in human lymphocytes. Clin. Immunol. Immunopathol. 71:14-18.
9) Nicoletti, I., G. Migliorati, M. C. Pagliacci, F. Grignani, and C. Riccardi. 1991. A rapid and simple method for measuring thymocyte apoptosis by propidium iodide staining and flow cytometry. J. Immunol. Methods. 139:271-279.
10) Ormerod, M. G., F. Paul, M. Cheetham, and X. M. Sun. 1995 Discrimination of apoptotic thymocytes by forward light scatter. Cytometry. 21:300-304.
11) Pandolfi, F., M. Pierdominici, A. Oliva, G. D'Offizi, I. Mezzaroma, B. Mollicone, A. Giovannetti, L. Rainaldi, I. Quinti, and F. Aiuti. 1995. Apoptosis-related mortality in vitro of mononuclear cells from patients with Hiv infection correlates with disease severity and progression. J. Acquir. Immune. Defic. Syndr. 9:450-458.
Solution Preparation Storage
Caution! PI is toxic, mutagen and carcinogen.
Appendix 2: Reagents
b) CO2 incubator, Heraeus Instruments GmbH (Hanau, Germany).
Figure 1-3. Cytograms produced by a typical analysis in a subject
with HIV-1 infection.

Fig. 2, left panel: simultaneous analysis of DNA content and
CD15 expression in all cells;
Fig. 2, right panel: analysis of DNA content and SSC in CD15-
cells (i.e. those present in R1 and R3).
Fig. 3: histogram referred to CD15- cells which shows a classical hypodiploic peak.