Analysis of Mitochondrial Membrane Potential
with the Sensitive Fluorescent Probe JC-1
 

Andrea Cossarizza and Stefano Salvioli
Department of Biomedical Sciences
University of Modena School of Medicine
via Campi 287, 41100 Modena, Italy
phone +39 59 428.613
fax +39 59 428.623
E mail: cossariz@unimo.it

 

 
Introduction



 

 

2. PROTOCOL


 
JC-1 staining

 

2.1 Materials
JC-1 fluorescent probe, plastic tubes for FACS analysis, complete medium, i.e. RPMI added with 10% fetal calf serum, PBS.
2.2. Methodology
1. Harvest cells (at least 2x105) from experimental samples, bring total volume up to 1 mL of fresh complete medium.

2. Stain cell suspension with 2.5 mg/mL JC-1. Shake cell suspension until the dye is well dissolved, giving a uniform red-violet color. To do this, it is also possible to vortex vigorously the suspension immediately after the addition of the probe.

3. Keep the samples in a dark place at room temperature for 15-20 minutes. The duration of the staining depends upon the cell type, but in our hands all the cells used (lymphocytes, cell lines of different origin, fibroblasts, keratinocytes, hepatocytes, etc.) responded quite well to the treatment. Wash twice centrifuging at 500 g for 5 min with a double volume of PBS.

4. Resuspend in 0.3 mL of PBS, then analyze immediatly with the flow cytometer, typically equipped with a 488 nm argon laser. Set the value of photomultiplier (PMT) detecting the signal in FL1 at about 390 V, and FL2 PMT at 320 V; FL1-FL2 compensation should be around 4.0%, while FL2-FL1 compensation around 10.6%. This is however the classical setting of the instrument we use in our laboratory, and it has to be taken into account that, as each instrument has a different sensitivity, a different setting can be necessary to obtain an optimal signal. Concerning instruments, the staining has been tested on several different apparatus such as an Excel, from Coulter (in Bergen, Norway), an Elite (Coulter) in Paris, some FACSCAN, a FACSTAR Plus and a FACSCalibur, from Becton Dickinson (in Krakow, Poland, or Modena and Venice, Italy), a Biorad Brite and a Partec (in Krakow too), and they work perfectly as well. Obviously, compensations have to be set in a different way.

 
3. COMMENTARY

 

3.1 Background information
The technique of JC-1 staining has been developed with the intent to detect DY in intact, viable cells. For this purpose JC-1 acts as a marker of mitochondrial activity, since the formation of J-aggregates, which give red emission, is reversible. Cells with high DY are those forming J-aggregates, thus showing high red fluorescence. On the other hand, cells with low DY are those in which JC-1 maintains (or re-acquire) monomeric form, thus showing only green fluorescence. Normally green fluorescence of depolarized cells is a little bit higher than that of polarized ones simply because of the presence of a higher amount of JC-1 monomers.

 
 

  1. Kroemer G., Zamzani N., Susin S.A. Mitochondrial control of apoptosis. Immunol. Today, 18: 44-51, 1997.

2. Susin S.A., Zamzami N., Castedo M., Daugas E., Wang H.G., Geley S., Fassy F., Reed J.C., Kroemer G. The central executioner of apoptosis: multiple connections between protease activation and mitochondria in Fas/APO-1/CD95- and ceramide-induced apoptosis. J. Exp. Med., 186: 25-37, 1997.

3. Kroemer G. The proto-oncogene Bcl-2 and its role in regulating apoptosis. Nature Med., 3: 614-620, 1997.

4. Cossarizza A., Kalashnikova G., Grassilli E., Chiappelli F., Salvioli S., Capri M., Barbieri D., Troiano L., Monti D., Franceschi C. Mitochondrial modifications during rat thymocyte apoptosis: a study at the single cell level. Exp. Cell Res., 214: 323-330, 1994.

5. Richter C., Schweizer M., Cossarizza A., Franceschi C. Control of apoptosis by the cellular ATP level. FEBS Lett., 378: 107-110, 1996.

6. Gorman A.M., Samali A., McGowan A.J., Cotter T.G. Use fo flow cytometry techniques in studying mechanisms of apoptosis in leukemic cells. Cytometry, 29: 97-105, 1997.

7. Yang J., Liu X., Bhalla K., Kim C.N., Ibrado A.M., Cai J., Peng T.I., Jones D.P., Wang X. Prevention of apoptosis by bcl-2: release of cytochrome c from mitochondria blocked. Science, 275: 1129-1132, 1997.

8. De Maria R., Lenti L., Malisan F., d'Agostino F., Tomassini B., Zeuner A., Rippo M.R., Testi R. Requirement for GD3 ganglioside in CD95- and ceramide-induced apoptosis. Science, 277: 1652-1654, 1997.

9. Maftah A., Petit J.M., Ratinaud M.H.A.J., R. 10-N nonyl-acridine orange: a fluorescent probe which stains mitochondria independently of their energetic state. Biochem. Biophys. Res. Commun., 164: 185-190, 1989.

10. Lopez-Mediavilla C., Orfao A., Gonzales M., Medina J.M. Identification by flow cytometry of two distinct rhodamine-123-stained mitochondrial populations in rat liver. FEBS Lett., 254: 115-120, 1989.

11. Terasaki M., Song J., Wong J.R., Weiss M.J., Chen B.L. Localization of endoplasmic reticulum in living and glutaraldehyde-fixed cells with fluorescent dyes. Cell, 38: 101-108, 1984.

12. Salvioli S., Ardizzoni A., Franceschi C., Cossarizza A. JC-1, but not DiOC6(3) or rhodamine 123, is a reliable fluorescent probe to assess DY in intact cells. Implications for studies on mitochondrial functionality during apoptosis. FEBS Lett., 411: 77-82, 1997.

13. Cossarizza A., Baccarani Contri M., Kalashnikova G., Franceschi C. A new method for the cytofluorimetric analysis of mitochondrial membrane potential using the J-aggregate forming lipophilic cation 5,5',6,6'-tetrachloro-1,1',3,3'-tetraethylbenzimidazolcarbocyanine iodide (JC-1). Biochem. Biophys. Res. Commun., 197: 40-45, 1993.

14. Cossarizza A., Salvioli S., Franceschini M.G., Kalashnikova G., Barbieri D., Monti D., Grassilli E., Tropea F., Troiano L., Franceschi C. Mitochondria and apoptosis: a cytofluorimetric approach. Fund. Clin. Immunol., 3: 67-68, 1995.

15. Cossarizza A., Ceccarelli D., Masini A. Functional heterogeneity of isolated mitochondrial population revealed by cytofluorimetric analysis at the single organelle level. Exp. Cell Res., 222: 84-94, 1996.

16. Hada H., Honda C., Tanemura H. Spectroscopic study on the J-aggregate of cyanine dyes. I. Spectral changes of UV bands concerned with J-aggregate formation. Photogr. Sci. Eng., 21: 83-91, 1977.

17. Reers M., Smith T.W., Chen L.B. J-aggregate formation of a carbocyanine as a quantitative fluorescent indicator of membrane potential. Biochemistry, 30: 4480-4486, 1991.

18. Smiley S.T., Reers M., Mottola-Hartshorn C., Lin M., Chen A., Smith T.W., Steele G.D., Chen L.B. Intracellular heterogeneity in mitochondrial membrane potential revealed by a J-aggregate-forming lipophilic cation JC-1. Proc. Natl. Acad. Sci. USA, 88: 3671-3675, 1991.

19. Johnson L.V., Walsh M.L., Bockus B.J., Chen L.B. Monitoring of relative mitochondrial membrane potential in living cells by fluorescence microscopy. J. Cell Biol., 88: 526-535, 1981.

20. Goldstein S., Korczack L.B. Status of mitochondria in living human fibroblasts during growth and senescence in vitro: use of the laser dye rhodamine 123. J. Cell Biol., 91: 392-398, 1981.

21. Darzynkiewicz Z., Staiano-Coico L., Melamed M.R. Increased mitochondrial uptake of rhodamine 123 during lymphocyte stimulation. Proc. Natl. Acad. Sci. USA, 78: 2383-2387, 1981.

22. Petit P.X., Lecoeur H., Zorn E., Dauguet C., Mignotte B., Gougeon M.-L. Alterations in mitochondrial structure and function are early events of dexamethasone-induced thymocyte apoptosis. J. Cell Biol., 130: 157-167, 1995.

23. Jenssen H.-L., Redmann K., Mix E. Flow cytometric estimation of transmembrane potential of macrophages - A comparison with microelectrode measurements. Cytometry, 7: 339-346, 1986.

24. Petit P.X., O'Connor D., Grunwald D., Brown S.C. Analysis of the membrane potential of rat- and mouse-liver mitochondria by flow cytometry and possible applications. Eur. J. Biochem., 194: 389-397, 1990.

 

 

 

Appendix 1: Stock solution:

 

JC-1 is dissolved in N,Ní-dimethylformamide (Sigma-Aldrich, cat. n. D8654) at the concentration of 2.5 mg/ml.

It is stored at -20°C. Light sensitive.

 

Appendix 2: Reagents

 
 
JC-1
Molecular Probes,
Eugene, OR, USA
catalog No.: T-3168
 

Note: colture medium, saline solutions and washing buffers are depending on the cell type which is used for the experimental procedure (PBMC, fibroblasts, hepatocytes, etc.). For blood white cells, RPMI 1640 with 10% heat inactivated foetal calf serum, 100 IU/ml penicillin, 100 mg/ml streptomycin, 2 mM L-glutamine is normally used as complete colture medium.

 

Appendix 3: Equipment

 
Flow Cabinet TC60 Gelaire 
Flow Cytometer FACScan Becton Dickinson
Incubator CO2-AUTO-ZERO Heraeus 
Centrifuge Minifuge RF  Heraeus
Pipetman P20, P200, P1000 Gilson
Vortex Vibrofix VF1 Electronic Janke & Kunkel-Ika 
Labortechnik
 

Appendix 4: Glossary

 

Mitochondrial membrane potential (Dy) is generated by mitochondrial electron transport chain, which drives a proton flow from matrix through inner mitochondrial membrane to cytoplasm, thus creating an electrochemical gradient. This gradient is in turn responsible for the formation of ATP molecules by F0-F1 ATP synthase. For this reason Dy is an important parameter for mitochondrial functionality and an indirect evidence of energy status of the cell.