This material was originally published in the Purdue Cytometry CD-ROM Series,volume 4

MULTIPARAMETRIC ANALYSIS
BY FLOW CYTOMETRY
Sandra Nuti
Chiron Vaccines Research Center, Siena, Italy
EMail: nuti@iris02.biocine.it
 

 1. Introduction
  2A. First protocol

 

2A.1 Materials

 2A.2 Methodology 1. Stain 106 cells into a FACS tube using a FITC-conjugated monoclonal antibody, performing a routinary suface staining method.

2. After 30 min of incubation on ice in the dark, wash by adding 3 ml of PBS-0.1% BSA and centrifuge at 250xg for 10 min at 4°C.

3. Fix the cells adding dropwise to the pellet 600 ml of pre-chilled (-20°C) 70% ethanol while gently vortexing. Incubate for 30 min at -20°C. The incubation can be performed also for longer periods (up to 24 hours). Centrifuge at 560xg for 5 min at 4°C.

4. Wash the pellet with 3 ml of PBS and centrifuge at 250xg for 10 min at RT.

5. Resuspend the pellet into 90 ml of PBS and add 10 ml of RNAse. Incubate for 30 min at RT in the dark, then add 150 ml of PI.

 

CAUTION: After each washing step remove very carefully the supernatant. Any remaining liquid on the top of the tube must be removed with absorbent paper.

 

2A.3. FACS Acquisition and Analysis

Figure 1: Example of a CD3 surface staining performed together with propidium iodide staining to detect apoptosis occurring in CD3+ T cells of a mouse spleen cells population. A mouse spleen cell population was analysed for apoptosis on CD3+ cells. The events were visualised using simultaneously the forward scatter (FSC) and the side scatter (SSC) parameters. Debris and dead cells were excluded by setting the FSC threshold and, during the analysis of the data, a region was drawn on viable cells (R1). Figure 1A shows the staining of the cells with an anti-CD3 FITC conjugated antibody by standard immunofluorescence techniques. In Figure 1A a region 2 (R2) has been created in the FL1 logarithmic scale on the CD3+ cells, and after combination of R1 (viable cells) and R2 (CD3+ cells) the DNA content analysis was performed in the FL2 linear scale. Figure 1B shows that apoptotic CD3+ cells are visualised as a subdiploid peak due to DNA fragmentation.

 2B. Second protocol

2B.1. Materials

Streptavidin-Phycoerythrin (SA-PE) (A2)

 

2B.2. Methodology

1. Stain 106 cells into a FACS tube using a FITC-conjugated monoclonal antibody, performing a routinary surface staining method.

2. After 30 min of incubation on ice in the dark wash by adding 3 ml of PBS-0.1%BSA and centrifuge at 250xg for 10 min at 4°C.

3. Fix the cellular pellet by adding, while gently vortexing, 450 ml of PBS and 150 ml of 4% PFA and incubate for 15 min at RT. Then add 3 ml of PBS-0.1% BSA and centrifuge at 250xg for 10 min.

4. Repeat the fixation adding dropwise to the pellet 600 ml of pre-chilled (-20°C) 70% ethanol while gently vortexing. Incubate for 30 minutes at -20°C. The incubation can be performed also for longer periods (up to 24 hours). Centrifuge at 560xg for 5 min at 4°C.

5. Wash the pellet with 3 ml of PBS-0.1% BSA and centrifuge at 250xg for 10 min at 4°C.

6. Add to the pellet 2 ml of 1x TdT buffer and centrifuge at 250xg for 10 min at 4°C.

7. To perform the TdT reaction, resuspend each sample in 40 ml of TdT buffer containing 100 mg/ml of BSA*, 0.3 nmoles of b-dUTP and 0.1 U/ml of TdT. Incubate for 45 min at 37°C in a water bath .

8. Add 3 ml of PBS-0.1% BSA and centrifuge at 250xg for 10 min at RT.

9. Resuspend the pellet into 90 ml of staining buffer and add 10 ml of SA-PE pre-diluted 1:10 into the staining buffer. Incubate for 30 min at RT in the dark and wash with 3 ml of PBS-0.1% BSA by centrifuging at 250xg for 5 min.

 

CAUTION: After each washing step remove the supernatant very carefully. Any remaining liquid on the top of the tube must be removed with absorbent paper.

 

2B.3. FACS Acquisition and Analysis
 

3. Commentary

 
3.1 Troubleshooting

During the methodology it may be possible to lose the cells. It is very important to use PBS-0.1% BSA for washing, and to follow carefully the speed and the time of the spinning steps.
 

1. Vaux, D. L., Weissman, I. L. and Kim, S. K. 1992. Prevention of programmed cell death in Caenorhabditis elegans by human bcl-2. Science 258: 1955.

2. Nicoletti, I., Migliorati, G., Pagliacci, M. C., Grignani, F. and Riccardi, C 1991. A rapid and simple method for measuring thymocyte apoptosis by propidium iodide staining and flow cytometry. J. Immunol. Meth. 139: 271.

3. Yamada, T. and Ohyama, H. 1988. Radiation-induced interphase death of rat thymocytes is internally programmed (apoptosis). Int. J. Radiat. Biol. 53: 65.

4. Williams, G. T., Smith, C. A., Spooncer, E., Dexter, T. N. and Tylor, D. R. 1990. Haemopoietic colony stimulating factors promote cell survival by suppressing apoptosis. Nature 343: 76.

5. Odaka, C., Kizaki, H. and Tadakuma, T. 1990. TCR-mediated DNA fragmentation and cell death in the T cell hybridoma. J. Immunol.144: 2096.

6. Arends, M. J., Morris, R. G. and Wyllie, H. 1990. Apoptosis: the role of endonuclease. Am. J. Pathol. 136: 593.

7. Swat, W., Ignatowicz, L. and Kisielow, P. 1991. Detection of apoptosis of immature Cd4+8+ thymocytes by flow cytometry. J. Immunol. Meth. 137: 79.

8. Gorczyca, W., Gong, J. and Darzynkiewicz, Z. 1993. Detection of DNA strand breaks in induvidual apoptotic cells by the in situ terminal deoxynucleotidyl transferase and nick translation assays. Cancer Res. 53: 1945.

 

Appendix 1 (A1): Stock solutions

 
 

Solution
Preparation
Storage 
Propidium iodide (PI) Prepare solution of 100 mg/ml PI in PBS or of 50 mg/ml PI in 0.1% sodium citrate plus 0.1% Triton X-100 (2). Store at 4°C for up to 1 week.
Paraformaldehyde 4% in PBS To prepare 100 ml of fixative: 

add 4 g of paraformaldehyde to 50 ml of water. Stir at 65°C on an hot plate (the solution remains slightly cloudy). Add 6 ml of 10 M NaOH. The solution will become clear. Filter trough a paper filter. Add 10 ml of 10x PBS. Adjust the volume to 100 ml with water. 

Store at 4°C for up to 24 hr or store aliquoted at -20°C.
10x TdT buffer To prepare 1 liter of 10x TdT buffer weigh: 138 gr cacodylic acid (100 mM final), 0.476 g cobalt chloride hexahydrate (0.2 mM final) and 0.154 g 1,4-dithio-DL-threitol (DTT) (0.1 mM final). Fill with distilled water to 1 liter and adjust the pH to 6.8 using NaOH tablets. Store aliquoted at -20°C.
Staining buffer 

 

To prepare 0.5 ml, mix: 100 ml 20x SSC, 50 ml Triton X-100 1%, 350 ml distilled H2O. Sterilize by auto- claving and store at RT.
20x SSC Dissolve 175.3 g NaCl and 88.2 g sodium citrate in 800 ml of H2O. Adjust the pH to 7.0 with few drops of a 10 N solution of NaOH. Adjust the volume to 1 liter with water.  Sterilize by auto- claving and store at RT.
 
Appendix 2 (A2): Reagents

Biotin-21-dUTP, 0.5 mmoles/100 µl 5021-1
Clontech

Bovine serum albumin A-9647
Sigma

Bovine serum albumin 711 454
for molecular biology, 20 mg/ml
Boehringer Mannheim

Cacodylic acid C-0125
Sigma

Cobalt chloride C-2644
Sigma

Paraformaldehyde 76240
Fluka

Propidium iodide P-4170
Sigma

RNAse Type I-A R-4875
Sigma

Sodium citrate S-4641
Sigma

Streptavidin-Phycoerythrin, 100 µg/0.4 ml 19539-014
GIBCO BRL

TdT enzyme, 300 U/15 µl M187/1,2
Promega

Triton X-100 T-9284
Sigma

Appendix 3 (A3): Equipment

Centrifuge mod. GS-6R
Beckman

Cytofluorimeter mod. FACScan or Vantage
Becton-Dickinson

Water bath mod. 12B
Julabo